Hi All:
I'm processing GCxGC data obtained from a Q Exactive GC instrument and the MS binning options does not seem to be working. The accurate masses are displayed only with one decimal (e.g., 121.3). I have tried the 'Uniform Bin' with many bin sizes and the spectrum exhibited in the 'Spectrum' window is the same with all bin sizes.
I'm using the v2020r2 GCxGC-HRMS and processing the CDF file.
Thanks,
Leo
Q Exactive GCxGC data
-
- Posts: 233
- Joined: October 7th, 2011, 10:16 am
Re: Q Exactive GCxGC data
There is a separate display setting in MS View. To change that setting,
Right click on the m/z column header on the Spectral Peak Table > adjust the decimal places to be displayed
, as shown below:
Right click on the m/z column header on the Spectral Peak Table > adjust the decimal places to be displayed
, as shown below: