Reextracted ion current
-
- Posts: 2
- Joined: November 12th, 2013, 6:53 am
Reextracted ion current
Hello, I´m starting with GC Image software and I use it for LC-MS data. I need obtain reextracted ion current (RIC) for example for m/z 184, how can I do it? can anybody help me, please? Thanks.
Last edited by Ovcacikova on November 25th, 2013, 7:27 am, edited 1 time in total.
-
- Posts: 236
- Joined: October 7th, 2011, 10:16 am
Re: Reextracted ion current
To view SIC, you need to:
View a spectrum in MS mode
-> In MS Viewer, Select a m/z range or enter the range in Edit Selected Range, for example, Max=184, Min=184.
-> Click on the Open SIC on the toolbar
Please see here for more details:
http://www.gcimage.com/lcxlc/usersguide ... tml#faqd.8
View a spectrum in MS mode
-> In MS Viewer, Select a m/z range or enter the range in Edit Selected Range, for example, Max=184, Min=184.
-> Click on the Open SIC on the toolbar
Please see here for more details:
http://www.gcimage.com/lcxlc/usersguide ... tml#faqd.8
-
- Posts: 2
- Joined: November 12th, 2013, 6:53 am
Re: Reextracted ion current
Thank you for your advice.