lnicolot
Posts: 14
Joined: January 23rd, 2014, 3:39 am

Internal Standard Quantitation

Post by lnicolot »

Hallo,
I am working using SIDA (Standard Isotope Dilution Analysis), is it possible to perform the "internal standard quantitation" using the ratio between qualifiuer and quantifier instead of "volume" and "1st quantifier" responses?
qtao
Posts: 209
Joined: October 7th, 2011, 10:16 am

Re: Internal Standard Quantitation

Post by qtao »

No, not right now although we are planning to support isotope dilution in the future. Currently a possible workaround is either to obtain the calibration function externally and then compute the quantities with Analysis CLIC of a blob, or use an area object overlapping with a blob. Both are a little bit tricky.
sarakhan
Posts: 1
Joined: October 31st, 2014, 1:58 am

Re: Internal Standard Quantitation

Post by sarakhan »

Is it possible to visualize more than 4 quantifiers or qualifiers in the blob table?
Save your time and get on 70-417 exam questions and answers in wikipedia and testking University of Wisconsin - Madison exams by using our latest CHECKPOINT and other superb exam pass resources of examsheets and IBM
qtao
Posts: 209
Joined: October 7th, 2011, 10:16 am

Re: Internal Standard Quantitation

Post by qtao »

You can add CLIC columns to show more quantifiers and qualifiers. For example, CLIC "Quantifier[5]" will show the 5th quantifier.

Please see here for details on how to add a CLIC column to Blob Table:
http://www.gcimage.com/gcxgc/usersguide ... lobalClics
AZhou
Posts: 1
Joined: February 24th, 2015, 8:32 am

Re: Internal Standard Quantitation

Post by AZhou »

Hello,

I have the similar problem with SIDA setup. You mentioned that it is doable by "using an area object overlapping with a blob". Could you give some detail instructions about the procedures? Thanks so much.

azhou
qtao
Posts: 209
Joined: October 7th, 2011, 10:16 am

Re: Internal Standard Quantitation

Post by qtao »

For R2.5, the area-object work-around is :

1. Create an area object overlapping with the blob
(http://www.gcimage.com/gcxgc/usersguide ... sFromBlobs)

2. Set the blob as ISTD

3. Specify different names for the blob and area object, for example, Blob is "Benzene-ISTD", Area is "Benzene".

4. Specify quantifier ions for the area object and the blob
(http://www.gcimage.com/gcxgc/usersguide ... #qionTable)

5. Quantify with the area object as usual
(http://www.gcimage.com/gcxgc/usersguide ... l#quantIon)

Note that the above workaround may or may not work for you, which depends on the calibration/quantitation equations you want to use.